Automatic test system switching hardware allows for high-speed testing of multiple devices under test, and is designed to reduce both test errors and costs. Designing a test system’s switching configuration requires an understanding of both the signals to be switched and the tests to be performed.

Switch Relays

The simplest definition of a switch is “a device that opens or closes a circuit.”

A relay
Relay
A relay is an electrically operated switch. Many relays use an electromagnet to operate a switching mechanism mechanically, but other operating principles are also used. Relays are used where it is necessary to control a circuit by a low-power signal , or where several circuits must be controlled...

 is an electronically operated switch. Three relay types are commonly used in automated test system switching:
  • Electromechanical relays are the most-often-used type because they have the largest signal range capability of the three. They also have the slowest actuation time and the shortest useful life. Electromechanical relays are ideal for high-voltage, high-current, and RF applications. A latching feature makes this relay type very appropriate for low-voltage applications where contact potential can interfere with the measurement.

  • Reed relays have actuation times of 0.5 ms to 2 ms and long life. By design, reed relays can only handle a portion of the signal range that electromechanical relays can offer. But they do provide a useful trade-off between increasing speed and maintaining signal integrity.

  • Solid-state relays can switch the fastest and have essentially infinite life. However, they can only handle small signal ranges and suffer from high on-resistance and high offset currents in the nanoamp range compared to picoamps of offset current for the other two types.

Switching’s Effect on Test System Accuracy

The ideal switch:
  • has no current limit during its ON state

  • has infinity resistance during its OFF state

  • has no voltage drop across the switch during its ON state

  • has no voltage limit during its OFF state

  • has zero rise time and fall time during state changes

  • switches only once without “bouncing” between on and off positions


It’s important to recognize, however, that real-life switches are not ideal, so when calculating the overall system accuracy, the effects of the switch itself and all the switching hardware in the system must be factored in.

As a signal travels from its source to its destination, it encounters various forms of interference and sources of error. Whenever a signal passes through a connecting cable or switch point, it may be degraded. For example, in low current and high resistance applications, unguarded cabling can introduce leakage current that will degrade measurement accuracy. Unshielded cable typically results in noisy readings for low current and high resistance applications.

Switch Relay Terminology

Three terms are used to describe the configuration of a relay: pole, throw, and form.
Pole refers to the number of common terminals within a given switch. Throw refers to the number of positions in which the switch may be placed to create a signal path or connection. Figure lA illustrates a single-pole, single-throw normally open switch (SPST NO). Figure 1B shows a single-pole, double-throw (SPDT) switch. One terminal is normally-open (NO) and the other is normally-closed (NC). Depending on the state of the switch, one or the other position is connected to the common terminal (COM). One signal path is broken before the other is connected, so this is referred to as a break-before-make configuration.
When more than one common terminal is used, the number of poles increases. Figure 1C shows a double-pole, single-throw (DPST) switch. Both poles are actuated simultaneously when the relay is energized. In this case, both poles are either always closed or always open. Figure 1D illustrates a double-pole, double throw (DPDT) switch.

Contact form, or simply form, is the term relay manufacturers use to describe a relay's contact configuration. "Form A” refers to a single-pole, normally open switch. "Form B" indicates a single-throw, normally closed switch, and "Form C" indicates a single-pole, double-throw switch. Virtually any contact configuration can be described using this format.

Switch System Topology

A variety of switching configurations are commercially available for test system development:

Scanner

The scanner (Figure 2) is used to connect multiple inputs to a single output in sequential order. Only one relay is closed at any one time. In its most basic form, relay closure proceeds from the first channel to the last, but some scanner systems allow skipping channels. Typical scanner switching applications include component burn-in testing, monitoring time and temperature drift in circuits, and taking data on system variables like temperature, pressure, flow, etc.

Multiplexer

Like a scan configuration, multiplex switching can be used to connect one instrument to multiple devices (1:N) or multiple instruments to a single device (N:1), but it offers much more flexibility than the scanner configuration because it permits multiple simultaneous connections and both sequential and non-sequential switch closures. Typical applications of multiplex switching include capacitor leakage, insulation resistance, and contact resistance test systems for multiple devices.

Matrix

The matrix switch configuration is the most versatile because it can connect multiple inputs to multiple outputs. A matrix is useful when connections must be made between several signal sources and a multipin device, such as an integrated circuit or resistor network.

Using a matrix switch card allows connecting any input to any output by closing the switch at the intersection (crosspoint) of a given row and column. The most common terminology to describe the matrix size is M rows by N columns (MxN). Matrix switch cards generally have two or three poles per crosspoint. As shown in Figure 3, a 5VDC source can be connected to any two terminals of the device under test (DUT). A function generator supplies pulses between another two terminals. Operation of the DUT can be verified by connecting an oscilloscope between two other terminals. The DUT pin connections can easily be programmed, so this system can be used to test a variety of components.
Some performance tradeoffs are typically necessary when choosing a matrix card for use with mixed signals. For example, if both high frequency and low current signals must be switched, take extra care when reviewing the specifications of the card. The card chosen must have wide bandwidth, as well as good isolation and low offset current. A single matrix card may not satisfy both requirements completely, so the system builder must decide which switched signal is more critical.

In a system with multiple cards, card types should not be mixed if their outputs are connected together. For example, a general-purpose matrix card with its output connected in parallel with a low current matrix card will degrade the performance of the low current card.

Expanding a Matrix

A large test system may require more rows and/or columns than a single switch card can accommodate, but it’s possible to expand a matrix by joining the rows and/or columns of several cards together. Depending on the switch card and mainframe selected, the rows of the cards may be connected together through the backplane of the mainframe or the rows may be wired externally.

Isolated switching

The isolated, or independent, switch configuration consists of individual relays, often with multiple poles, with no connections between relays. Isolated relays are commonly used in power and control applications to open and close different parts of a circuit that are at substantially different voltage levels. Applications for isolated relays include controlling power supplies, turning on motors and annunciator lamps, and actuating pneumatic or hydraulic valves. Figure 4 illustrates a single isolated relay or actuator, in which a single-pole normally open relay is controlling the connection of the voltage source to the lamp. This relay connects one input to one output. An isolated relay can have more than one pole and can have normally closed contacts as well as normally open contacts.
Isolated relays are not connected to any other circuit, so the addition of some external wiring makes them suitable for building very flexible and unique combinations of input/output configurations.

Given that the relays are isolated from each other, the terminals of each channel on the switch card are independent from the terminals of the other channels. As shown in Figure 5, each isolated Form A relay has two terminals. Two-pole isolated relays would have four terminals (two inputs and two outputs). A Form C isolated relay would have three terminals.

Cold vs. Hot Switching

The term cold switching indicates that a switch is activated with no signal applied. Therefore, no current will flow when the switch is closed, and no current will be interrupted when the switch is opened. In contrast, in hot switching, voltage is present and current will flow immediately once the contacts close. When the switch is opened, this current will be interrupted and can cause arcing.

Cold switching allows power to be applied to the device under test in a controlled manner. Its primary advantage is a longer switch life than with hot switching (up to a thousand times more cycles than with hot switching). Cold switching also eliminates arcing at the relay contacts and any radio frequency interference that might be caused by arcing. Hot switching might be necessary if close control must be exercised in the period between the application of power and the making of the measurement. For example, hot switching is typically used where digital logic is involved, because devices might change state if power is interrupted even for a moment.

With relatively large relays, hot switching may be necessary to ensure good contact closure. The connection may not be reliable without the “wetting” action of the current through the contacts.